Well, Mechanikus visited me today with his fine timepieces and we tested them all on my Witschi at room temperature of 23 degrees of Celsius:
The Citizen (Cal.A610H):
- inhibition period: 10 seconds
- accuracy of the stepping motor (by the magnetic sensor): -0.01 s/d
- accuracy of the quartz oscillator (by the acoustic sensor):
1st second: +7.54 s/d
2nd second: +7.50 s/d
3rd second: +7.54 s/d
4th second: +7.25 s/d
5th second: +6.17 s/d
6th second: +6.09 s/d
7th second: +6.14 s/d
8th second: +6.35 s/d
9th second: +6.19 s/d
10th second: +6.19 s/d
then it started again: the results might have changed a little bit here and there but the trend remained the same for the next lot of 10 seconds then the following lot of 10... and so on... and that is very similar to the test results of my thermocompensated Citizen Exceed watches: continuous inhibition (digital count suppression) combined with the continuous adjustment of the frequency of the quartz oscillator(!).
Citizen Crystron 4 Mega (Cal.7370D):
- inhibition period: none
- accuracy of the stepping motor (by the magnetic sensor): -0.03 s/d
- accuracy of the quartz oscillator (by the acoustic sensor): could not be measured
(note: the Witschi QT6000 is designed to measure 32kHz quartz oscillators only by its acoustic sensor)
Longines Flagship VHP Perpetual Calendar (L.546.2 = ETA 252.611):
- inhibition period: 8 minutes (480 seconds)
- accuracy of the stepping motor (by the magnetic sensor): -0.00 s/d
- accuracy of the quartz oscillator (by the acoustic sensor): between +2.70s/d and +2.90 s/d
Seiko Perpetual Calendar (Cal.8F32):
- inhibition period: 10 seconds
- accuracy of the stepping motor (by the magnetic sensor): +0.23 s/d
- accuracy of the quartz oscillator (by the acoustic sensor): could not be measured
(note: the Witschi QT6000 is designed to measure 32kHz quartz oscillators only by its acoustic sensor)
Junghans Radio-controlled analog quartz:
- inhibition period: none
- accuracy of the stepping motor (by the magnetic sensor): -0.08 s/d
- accuracy of the quartz oscillator (by the acoustic sensor): -0.08 s/d
Seiko Spring-Drive:
Mechanikus observed an accuracy of around +10 s/month over a longer period. The Witschi failed to get proper reading from the watch though we tried many times. Well, actually once we got a result of +0.34 s/d for the stepping motor by the magnetic sensor using 4 minutes (480 seconds) measuring period but we could not repeat it so we can't officially accept it though that result is very similar to Mechanikus' earlier observation.
Casio ProTrek Solar Atomic (my radio-controlled digital multifunction watch):
- inhibition period: none
- accuracy of the quartz oscillator (by the acoustic sensor): +0.50 s/d




LinkBack URL
About LinkBacks
Reply With Quote
) - I believe we see a generic 10 seconds inhibition used for raw rate correction coupled with a independent 2-frequency-variable-fill-factor TC method which seems to be a slightly more modern version of the method used in the older 27xx calibers!